[1] #Hemodynamics Tweetorial #2
Heart failure pt in ICU is -3.5L after 2d of aggressive diuresis. On day 3, urine output is ⬇️and BUN/Cr is ⬆️
You personally wedge #PAC at bedside and obtain a mean wedge 17 mmHg (a normal mean wedge is 6-10 mmHg). Admission wedge was 24
Heart failure pt in ICU is -3.5L after 2d of aggressive diuresis. On day 3, urine output is ⬇️and BUN/Cr is ⬆️
You personally wedge #PAC at bedside and obtain a mean wedge 17 mmHg (a normal mean wedge is 6-10 mmHg). Admission wedge was 24
[2]
Question 1. Based on wedge 17 mm Hg (good waveform; no trick), you:
Question 1. Based on wedge 17 mm Hg (good waveform; no trick), you:
[4]
🔹”Zeroing”
🔹”Leveling”
🔹”Damping”
🔹”Zeroing”
🔹”Leveling”
🔹”Damping”
[5]
🔹”Zeroing”
-Device is “zeroed” to open the air-fluid interface to atmospheric pressure (AP)
-AP is the reference standard against which other pressures are measured
-“Re-zeroing” is done often since transducer and atmospheric pressure drift away from calibration point
🔹”Zeroing”
-Device is “zeroed” to open the air-fluid interface to atmospheric pressure (AP)
-AP is the reference standard against which other pressures are measured
-“Re-zeroing” is done often since transducer and atmospheric pressure drift away from calibration point
[7]
To zero:
-Point the 3-way stopcock “off” towards the patient (points up on the transducer)
-This now allows the vent port to be “opened” to atmospheric pressure
-If vent port cap is not fenestrated, it has to be temporarily removed
(port cap in video 👇🏻 is fenestrated)
To zero:
-Point the 3-way stopcock “off” towards the patient (points up on the transducer)
-This now allows the vent port to be “opened” to atmospheric pressure
-If vent port cap is not fenestrated, it has to be temporarily removed
(port cap in video 👇🏻 is fenestrated)
[9]
🔹”Leveling”
-Selection of a point of interest (R atrium for #PAC) at which the ref standard (zero) is set
-RA is considered the phlebostatic axis (RA is also used as the ref point for CVP assessm)
*REMEMBER:
Pressure changes 7.4mmHg for every 10 cm transd is ⬆️ or ⬇️ RA
🔹”Leveling”
-Selection of a point of interest (R atrium for #PAC) at which the ref standard (zero) is set
-RA is considered the phlebostatic axis (RA is also used as the ref point for CVP assessm)
*REMEMBER:
Pressure changes 7.4mmHg for every 10 cm transd is ⬆️ or ⬇️ RA
[12]
Answer to Question 1 is: “Can I trust that wedge?”
In our case, when wedge 17 mmHg was obtained, transducer wasn’t leveled and pt’s RA was about 11 cm above the transducer.
Therefore, real wedge was <10 mmHg.
Diuresis was held for a few hrs and restarted at home dose
Answer to Question 1 is: “Can I trust that wedge?”
In our case, when wedge 17 mmHg was obtained, transducer wasn’t leveled and pt’s RA was about 11 cm above the transducer.
Therefore, real wedge was <10 mmHg.
Diuresis was held for a few hrs and restarted at home dose
BONUS:
If you see pressure tracings that become negative (below Zero), consider an unaddressed LEVELING issue before you interpret and use that #PAC info
See 👇🏻 RAP tracing: transducer was higher than RA level, causing sub-Zero values (RA level is ⬇️ = pressure is “falsely” ⬇️)
If you see pressure tracings that become negative (below Zero), consider an unaddressed LEVELING issue before you interpret and use that #PAC info
See 👇🏻 RAP tracing: transducer was higher than RA level, causing sub-Zero values (RA level is ⬇️ = pressure is “falsely” ⬇️)
[14]
Which answer is FALSE regarding an OVERDAMPED waveform
Which answer is FALSE regarding an OVERDAMPED waveform
[17]
Which answer is FALSE regarding an UNDERDAMPED waveform:
Which answer is FALSE regarding an UNDERDAMPED waveform:
[21]
TAKE-HOME Messages (applies to ICU and outpt cardiac/PH pts)
🔷Before you interpret/use #PAC info, verify:
🔸Zeroing (atmosp pressure)
🔸Leveling (match transducer level to RA, AKA phlebostatic axis)
🔸Waveforms (to exclude OVER or UNDER damping affecting actual values)
TAKE-HOME Messages (applies to ICU and outpt cardiac/PH pts)
🔷Before you interpret/use #PAC info, verify:
🔸Zeroing (atmosp pressure)
🔸Leveling (match transducer level to RA, AKA phlebostatic axis)
🔸Waveforms (to exclude OVER or UNDER damping affecting actual values)
TAKE-HOME Messages 2
🔸If you see -ve (below Zero) pressures, especially in non-ill pts, consider leveling error
🔸Over and Underdamping affect Syst and Diast pressures; MAP is usually preserved
🔸Verify line flushing/waveform every time a MVsat is drawn from distal #PAC port
🔸If you see -ve (below Zero) pressures, especially in non-ill pts, consider leveling error
🔸Over and Underdamping affect Syst and Diast pressures; MAP is usually preserved
🔸Verify line flushing/waveform every time a MVsat is drawn from distal #PAC port
Additional relevant reading:
-NEJM Art Line Pressure Transducer
nejm.org
-Deranged physiology
derangedphysiology.com
-NEJM Art Line Pressure Transducer
nejm.org
-Deranged physiology
derangedphysiology.com