[1] #Hemodynamics Tweetorial #2
Heart failure pt in ICU is -3.5L after 2d of aggressive diuresis. On day 3, urine output is ⬇️and BUN/Cr is ⬆️
You personally wedge #PAC at bedside and obtain a mean wedge 17 mmHg (a normal mean wedge is 6-10 mmHg). Admission wedge was 24
[2]
Question 1. Based on wedge 17 mm Hg (good waveform; no trick), you:
[3]
Before you use the pulmonary artery catheter #PAC information, you should know its pitfalls
Given the potential impact of #PAC-related decisions, consider learning as much (or more) as the ICU nurse/cath lab tech about #PAC setup and troubleshooting
[4]
🔹”Zeroing”
🔹”Leveling”
🔹”Damping”
[5]
🔹”Zeroing”
-Device is “zeroed” to open the air-fluid interface to atmospheric pressure (AP)
-AP is the reference standard against which other pressures are measured
-“Re-zeroing” is done often since transducer and atmospheric pressure drift away from calibration point
[6]
Transducer components and directions of the components 👇🏻
[7]
To zero:
-Point the 3-way stopcock “off” towards the patient (points up on the transducer)
-This now allows the vent port to be “opened” to atmospheric pressure
-If vent port cap is not fenestrated, it has to be temporarily removed
(port cap in video 👇🏻 is fenestrated)
[8]
🔹”Leveling”
Question 2. For #PAC, you place the transducer to match the level of:
[9]
🔹”Leveling”
-Selection of a point of interest (R atrium for #PAC) at which the ref standard (zero) is set
-RA is considered the phlebostatic axis (RA is also used as the ref point for CVP assessm)
*REMEMBER:
Pressure changes 7.4mmHg for every 10 cm transd is ⬆️ or ⬇️ RA
[10]
To level:
Identify the point at which the:
🔹mid-axillary line intersects with the
🔹4th intercostal space
That point should be at the same level of the transducer’s diaphragm
Consider using an accurate leveling system (e.g., laser) to match transducer ht with RA ht
[11]
Leveling is 🔑
🔷IN RELATIONSHIP TO TRANSDUCER:
🔹For every 10cm that RA “rises” above transducer, pressure falsely “rises” 7.4 mmHg
🔹For every 10cm that RA is “lowered” below transducer, pressure falsely “lowers” 7.4 mmHg

RA ⬆️ = pressure ⬆️
RA ⬇️= pressure ⬇️
[12]
Answer to Question 1 is: “Can I trust that wedge?”
In our case, when wedge 17 mmHg was obtained, transducer wasn’t leveled and pt’s RA was about 11 cm above the transducer.
Therefore, real wedge was <10 mmHg.
Diuresis was held for a few hrs and restarted at home dose
BONUS:
If you see pressure tracings that become negative (below Zero), consider an unaddressed LEVELING issue before you interpret and use that #PAC info
See 👇🏻 RAP tracing: transducer was higher than RA level, causing sub-Zero values (RA level is ⬇️ = pressure is “falsely” ⬇️)
[13]
🔹”Damping”
Damping is the “shock absorber” effect on the pressure line (like a car suspension)
Overdamping:
Sluggish oscillations
Underdamping:
Oscillations are too pronounced
Causes of Over and Under-Damping 👇🏻
[14]
Which answer is FALSE regarding an OVERDAMPED waveform
[15]
🔹OVER-damping
🔸Slurred upstroke
🔸ABSENT Dicrotic Notch
🔸Systolic is Underestimated
🔸Diastolic is Overestimated
🔸MAP is usually preserved
🔸Pulse pressure is underestimated (affects PAPi calc)
🔸The waveform gets “squished” (think letter “O” for “O”verdamped)
[16]
How to address an OVERDAMPED #PAC (and arterial) waveform
👇🏻
[17]
Which answer is FALSE regarding an UNDERDAMPED waveform:
[18]
🔹UNDER-damping
🔸Non-physiologic oscillations during Diast phase
🔸Deep dicrotic notch
🔸Systolic is Overestimated
🔸Diastolic is Underestimated
🔸Pulse pressure overestimation
🔸MAP is usually preserved
🔸Waveform gets “stretched out” (inverted “U” for “U”nderdamped”
[19]
How to address an UNDERDAMPED #PAC waveform
👇🏻
[20]
Besides looking at the waveforms, consider FAST-FLUSH test (square wave test) to assess for waveform damping characteristics
🔸Hold fast-flush device for <1 sec
🔸Assess the square wave and characteristics of the wave (s) that follow
See image 👇🏻
[21]
TAKE-HOME Messages (applies to ICU and outpt cardiac/PH pts)
🔷Before you interpret/use #PAC info, verify:
🔸Zeroing (atmosp pressure)
🔸Leveling (match transducer level to RA, AKA phlebostatic axis)
🔸Waveforms (to exclude OVER or UNDER damping affecting actual values)
TAKE-HOME Messages 2
🔸If you see -ve (below Zero) pressures, especially in non-ill pts, consider leveling error
🔸Over and Underdamping affect Syst and Diast pressures; MAP is usually preserved
🔸Verify line flushing/waveform every time a MVsat is drawn from distal #PAC port
Additional relevant reading:
-NEJM Art Line Pressure Transducer
nejm.org
-Deranged physiology
derangedphysiology.com